Inter-Run Calibration
You need to do inter-run calibration if you want to compare samples from different runs e.g.:
- when it is not possible to get all samples for the same gene on the same plate
- when you do additional runs weeks or months after your initial experiment
Of course there is a lot of variability between runs on a qPCR instrument:
- thermal block is not always heating uniformously
- quality of the lamp, the filters and the detector decreases over time
- data analysis settings on the qPCR instrument (baseline correction and threshold) can be slightly different
- efficiency of reagents (polymerase, fluorophores) is variable
- optical properties of the plastic plates vary
Fortunately, inter-run calibration allows you to eliminate most of this variability.
In this experiment we will analyze the data from the gene expression experiment (see Analyzing gene expression data in qbase+) together with data from 2 runs (Run4 and Run5) that were done weeks after the initial gene expression experiment.
Because the data comes from two different experiments spread over time, we have included three inter-run calibrators on the plates: Sample01, Sample02 and Sample03.
The principle of the IRCs is very similar to that of the reference genes: In theory, the IRCs should have the same NRQ in each run. In practice, the difference in NRQ between two runs is a measure of the inter-run variation and can be used to adjust the NRQs to remove the inter-run variation.